Power Supply
(CR-Series)¶
CR-series devices can be supplied with power via Power over Ethernet or via a 24 V supply on the GPIO port. If a DC supply is connected it will be preferred over PoE.
The camera can internally buffer energy for approximately 1.5 ms flash time at full output power, regardless of DC or PoE operation. After this period the actual output power will reduce to the power level available from the PoE or DC supply. If you need long flash times at full output power the camera should be powered via the DC supply.
Power-over-Ethernet (PoE)¶
On PoE switches the camera registers as 802.3bt (“PoE++”) device with PD class 8 (90 W). Although the device might power up on switches providing a lower PoE class, we currently only support operation on PoE++ equipment providing PD class 8 (90 W). This is necessary to provide enough power headroom for longer projector flash times. Depending on the switch/injector model, cable length and number of connected devices on the switch the camera projection power may nevertheless need to be limited in software for stable operation for long flash times.
Operation on PoE++ equipment providing less than PD class 8 (90 W) is unsupported.
DC Supply¶
The camera can be powered via a 24 V 10 A DC supply on the GPIO port. The full 10 A are only required for short periods during projector flashes. The device idle current is ~0.35 A. The average current will be below 1.4 A in all cases, due to LED duty cycle limits. For reduced power requirement the projection power can be reduced in software.
You should ensure at least 20.4 V during peak current operation at the device and at most 28.8 V maximum voltage when idle. When supplying increased voltage please make sure to leave a safety margin to the overvoltage protection (internal TVS diode) which will be triggered above 32 V. DC voltage drops below 20 V might result in reduced LED output power and the device could also intermittently switch to draw power from PoE if available.
The devices are equipped with an internal 4 A fuse with a mean opening time of 1100 ms at 10 A.
For protection of cabling against electrical faults use a power supply with 10 A peak power and maximum 5 A average power rating, or insert a slow blow fuse at the supply.
Long Cable Lengths¶
The high peak power requirement during projector flashes will limit the usable cable length. To cover longer distances and still supply sufficient voltage during peak power phases at the device input we recommend the following measures:
Use cabling with sufficient wire cross section for covering most of the distance and attach an adapter to M12 A-coded in front of the device.
You can supply up to 28 V when the device is idle, which allows for more losses while still keeping the 20 V minimum voltage level during peak power phases.
If these measures are not sufficient you can still limit peak power in software, which will come at a cost of slightly higher exposure times and slightly lower environmental light robustness.
Calculation Examples¶
The following table lists some examples of supply and wiring combinations for different distances.
Cable Length, Connector, Supply |
Wire Thickness |
Resistance, GND+Supply Wire |
Voltage Drop on Cable |
Min. Voltage at Device During Flash |
15 m, M12 |
1.5 mm² |
12mΩ / m * 15m * 2 = 0.36Ω |
0.36Ω * 10A = 3.6V |
24V–3.6V = 20.4V |
25 m, M12, 28 V supply |
1.5 mm² |
12mΩ / m * 25m * 2 = 0.60Ω |
0.60Ω * 10A = 6.0V |
28V–6.0V = 22.0V |
25 m, custom adapter |
2.5 mm² |
7mΩ / m * 25m * 2 = 0.35Ω |
0.35Ω * 10A = 3.5V |
24V–3.5V = 20.5V |
50 m, custom adapter, 28 V |
2.5 mm² |
7mΩ / m * 50m * 2 = 0.70Ω |
0.70Ω * 10A = 7.0V |
28V–7.0V = 21.0V |